tape tension sensor
Posted: Sat Feb 02, 2008 6:11 am
Hi to all and expecially to you Greg,
I'm trying to make some adjustment and calibration on the tape path of my MS16 Tascam atr and just made the measurements of the left and right tension' sensors output voltages: making a comparition to the reference tensions on manual I found some slight differences.
In Play, Stop and Edit mode the tensions are 3.73-3.75 on 3.7V reference; between left and right sensors there is difference of 0.2V
In Ffwd mode the reference would be 5.2V and i found 4.61 on left and 5.15 on right
In Rew mode reference would be 3.5V and here there are 4.9 on left and 4.6 on right.
My questions are:
1) in Ffwd and Rew the output voltage must be the same on left and right? or it refers to direction of the tape? (Ffwd= right sensor - Rew= left sensor)
2) in your experience is a 0.5V difference (from reference) so important?
Thanks very much
I'm trying to make some adjustment and calibration on the tape path of my MS16 Tascam atr and just made the measurements of the left and right tension' sensors output voltages: making a comparition to the reference tensions on manual I found some slight differences.
In Play, Stop and Edit mode the tensions are 3.73-3.75 on 3.7V reference; between left and right sensors there is difference of 0.2V
In Ffwd mode the reference would be 5.2V and i found 4.61 on left and 5.15 on right
In Rew mode reference would be 3.5V and here there are 4.9 on left and 4.6 on right.
My questions are:
1) in Ffwd and Rew the output voltage must be the same on left and right? or it refers to direction of the tape? (Ffwd= right sensor - Rew= left sensor)
2) in your experience is a 0.5V difference (from reference) so important?
Thanks very much